The "dry" signalling threshold is set simultaneously with the low level calibration. When the
conductivity measured here is undershot, the sensor signals the "dry" condition with a
current output value of 2.4 mA. Given different products with widely differing conductivity, the
4 mA adjustment should be carried out using the least conductive liquid.